Over 500 Total Lots Up For Auction at One Location - NJ 06/15

philips envisor c problem: unable to initailse patient data

(1)

(1)

suresh reddy

philips envisor c problem: unable to initailse patient data

May 22, 2008 12:36

i have a problem in philips envisor c [ ultrasound equepment]
problem: unable to initailse patient data, measurments, report
this makes me unable to do measurments, unable to retrive patient data

help required

Log inor Register

to rate and post a comment

(373)
(66)
(15)

yuan bo

Re: philips envisor c problem: unable to initailse patient data

May 22, 2008 03:25

I think maybe the hard disk is fault or the software have problem

Log inor Register

to rate and post a comment

(39)
(1)
(2)

Emerson Freire

Re: philips envisor c problem: unable to initailse patient data

May 22, 2008 09:16

You probably have a corrupted database.
To solve this you will need to reinstall the software.

Log inor Register

to rate and post a comment

Cam Muir

re: philips envisor c problem: unable to initailse patient data

May 30, 2014 08:39

I have the same problem .. did you find a solution?

Log inor Register