Philips Envisor Hd internal error

February 18, 2015 08:37

hello

My device is giving such an error. (Internal Error)

Needed to service password so I can control.

Can you help?


Philips Envisor HD version C.1.1

1 person liked this
Log inor Register

to rate and post a comment

(119)
(57)
(6)

Jim Carr

re: Philips Envisor Hd internal error

February 19, 2015 10:20

A password should not be needed and it will not clear the error, because it means something is broken. Try replacing the token ring cable, or check both ends of the token ring cable for bent pins. If bent pins are found, check the female connector for damage from the bent pin.

If the token ring cable is OK, try to go into simulation mode by pressing Ctrl+Alt+s on the keyboard. If the ultrasound application now starts and you see the ultrasound image area on the screen, the PC is OK. If it does not go into simulation mode, the PC could be the cause.

2 people liked this
Log inor Register

to rate and post a comment

(40)
(8)
(1)

Paul Galo

re: Philips Envisor Hd internal error

February 19, 2015 02:05

Please send an email to paulgalo@gmail.com to provide you the passwords

1 person liked this
Log inor Register

to rate and post a comment

(18)
(5)

Cagri GUZEL

re: re: Philips Envisor Hd internal error

February 20, 2015 10:19

cagriguzel@armed.com.tr

Thank you so much...

Log inor Register

to rate and post a comment

stats
DOTmed text ad

BK and other Urological Systems available. BK Probe repair available!!!

Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net

DOTmed Certified stats
advertisement

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)