Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Philips HD 11

Tags: Ultrasound

Aftab

Philips HD 11

February 03, 2014 06:39

Dear All,

I having a problem measurements Unavailable error system show following massage
The report package is initializing. Initialization completes approximately 30 sec after imaging starts (if the system has been moved to another network, initialization can take up 2 minutes)please try measurement again. please help me what i do.
regards,

engr sohail
email sohail_aftab@live.com

Log inor Register

to rate and post a comment

(119)
(57)
(6)

Jim Carr

re: Philips HD 11

February 05, 2014 07:56

If that is the only thing not working correctly, it sounds like corrupted software. I would try restoring the installed options from the backup CD. They are part of the "System Settings."