Hi everybody I have this issue: The Ultrasound machine write on the monitor an alert about High temp. But all the Fan works good so i think it's a sensor issue. Someone knows how and where the sensor works? Thanks
Could be the sensor. But, start with complete cleaning (open/clean, blow everything with compressed air, clean the filters and fans, check the air circuit).
Be sure that both sets of fans are working, the ones for the power supply and the ones for CPU station. Also check the sensor for debris around it. If the supply fans aren't working but the CPU's are it will detect this pretty quickly.
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
Pheolo
Philips HD11 High temperature
October 18, 2013 10:14
Hi everybody I have this issue: The Ultrasound machine write on the monitor an alert about High temp. But all the Fan works good so i think it's a sensor issue. Someone knows how and where the sensor works? Thanks
to rate and post a comment
Hossam Eldin Zidan
re: Philips HD11 High temperature
October 18, 2013 06:22
I have the same problem and I wish I can know the reason for this message, please let me know if you receive the correct answer, thank you.
to rate and post a comment
Kobris
re: Philips HD11 High temperature
October 19, 2013 08:28
Could be the sensor. But, start with complete cleaning (open/clean, blow everything with compressed air, clean the filters and fans, check the air circuit).
to rate and post a comment
Lobsang Ortiz
re: re: Philips HD11 High temperature
October 21, 2013 02:38
Sometimes, the micro needs a little grass, so, you can check it and put it a little device electronic grass.
to rate and post a comment
Zach Ruder
re: Philips HD11 High temperature
October 21, 2013 02:38
Be sure that both sets of fans are working, the ones for the power supply and the ones for CPU station. Also check the sensor for debris around it. If the supply fans aren't working but the CPU's are it will detect this pretty quickly.
to rate and post a comment