Darren Montano

Philips HD3

November 01, 2014 02:11

We do have a Philips HD3 Ultrasound machine that is showing unclear image on the monitor.
Is there anyone could extend help/advise?
Do we need to change the Transducer or Update the Operating system/system drivers.

Thanks in advance
Darren
red22.cls.afg@gmail.com

Log inor Register

to rate and post a comment

(168)
(40)

Kobris

re: Philips HD3

November 03, 2014 08:25

Hi
is the whole image blurry, everything or just ultrasonic beam? If whole, then the monitor is faulty.
If the beam is blurry -> I would bet that the transducer is old, the cristals got weaken...
I doubt the software upgrade would resolve it.
BR

Log inor Register

to rate and post a comment

Darren Montano

re: re: Philips HD3

November 03, 2014 03:40

thanks let me double check, I would try to use other transducer that is compatible to HD3, by any chance you have a Programming disc of HDI 4000 in where I have loaded the one I have got it says Error 0

Log inor Register

to rate and post a comment

Darren Montano

re: re: Philips HD3

November 03, 2014 03:41

Is it possible to clean or service the transducer?

Log inor Register

to rate and post a comment

(168)
(40)

Kobris

re: re: re: Philips HD3

November 03, 2014 06:46

if the cristals are weak then no.. try with another one

Log inor Register

to rate and post a comment

Darren Montano

re: re: re: re: Philips HD3

November 04, 2014 04:24

thanks ...more power to you