Derbal Abir
philips HDI 5000 CT
March 26, 2018 04:09
hi
i have a philips HDI 5000 sonoCT shows me the error message 008.089
i need the code error and service manual
some one can help me
to rate and post a comment
Rey Gonzalez
re: philips HDI 5000 CT
March 27, 2018 04:24
008 means configuration error, 089 means not comunication between APM and ultrasound, since you are geeting both i would say your APM/DVS module is not booting at all and the ultrasound it is saying i cant find all hardware and cant connect to APM/DVS. Your solution replace defective module or remove it and reprogram NVRams on CPU.
I hope it helps.