i have a philips HDI 5000 sonoCT shows me the error message 008.089
i need the code error and service manual
some one can help me

Log inor Register

to rate and post a comment

(230)
(35)
(4)

Rey Gonzalez

re: philips HDI 5000 CT

March 27, 2018 04:24

008 means configuration error, 089 means not comunication between APM and ultrasound, since you are geeting both i would say your APM/DVS module is not booting at all and the ultrasound it is saying i cant find all hardware and cant connect to APM/DVS. Your solution replace defective module or remove it and reprogram NVRams on CPU.
I hope it helps.