Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Philips HDI-4000 - need authorization code to set up DICOM

(21)
(2)
(2)

Ilia Davidov

Philips HDI-4000 - need authorization code to set up DICOM

December 11, 2008 02:56

Dear Sirs!
Please help me to get an authorization code (password) for my HDI-4000. I need to reconfigure IP address for DICOM - and it asks me for code. Can I reset the password to default if previous service engineer set custom one?Pleas help!:) Thanks!

Log inor Register

to rate and post a comment

(21)
(2)
(2)

Ilia Davidov

Re: Philips HDI-4000 - need authorization code to set up DICOM

December 17, 2008 01:36

Let me answer myself :) I've found an authorization key via debugger - it is 'bothell'. So if you need to get a network settings on HDI 4000 - just go to SETUP (Press SETUP button) then DICOM, then Network Settings (or Configuration - I'm not sure) - then enter the 'bothell' - and the standard W2K Network Settings window appears. By the way, it's an ordinary Explorer, so you can put any location (e.g. C:\) in address bar and work with 4000 as on simple PC. Hope this info will become usefull for somebody :)

Log inor Register

to rate and post a comment

(4)
(1)

Edson Barrios

Re: Philips HDI-4000 - need authorization code to set up DICOM

August 20, 2009 06:10

looking for service manual and oprator manual ATL hdi 4000.
Can you help??

Log inor Register

to rate and post a comment

(21)
(2)
(2)

Ilia Davidov

Re: Philips HDI-4000 - need authorization code to set up DICOM

August 21, 2009 11:25

Please give me your E-mail, I'll send the service manual to you.

Log inor Register

to rate and post a comment

(4)
(1)

Edson Barrios

Re: Philips HDI-4000 - need authorization code to set up DICO

August 21, 2009 05:36

Hi.
And very much thanks for helping.
I did not recive it yet. Please send it to:
dr.edson.barrios@gmail.com and or:
dr_edson_barrios@hotmail.com
Again thanks a lot.
Sincerely: Dr Edson Barrios.

Log inor Register

to rate and post a comment

raheel nawaz

Re: Philips HDI-4000 - need authorization code to set up DICO

October 11, 2011 08:48

can you plz send me a copy of the philips Atl hdi 4000 service mannual
nwassem@gmail.com iwill be great full
regards

Log inor Register

to rate and post a comment

(3)

Dan Hrituc

re: Re: Philips HDI-4000 - need authorization code to set up DICOM

November 29, 2012 02:55

Please send me the service manual for Philips HDI 4000 if you have.Thank you very much.
My e-mail dhritzuk@medical-electronic.ro

Log inor Register

to rate and post a comment

(45)
(8)

carlos enrique salazar gea

re: Re: Philips HDI-4000 - need authorization code to set up DICOM

May 07, 2014 09:47

Could you send me a copy¿
csalazargea@hotmail.com
My equipment no booting.
Regards

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Medical Trailer, Transport & Storage. 40 years experience.

One call to 440-263-6575 and your move will be covered. We repair & store trailers. We build to your custom needs. JDI Mobile Medical Manufacturing. Inc. We have your back.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)