If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare
Samantha Legeny
Philips IDI 5000 error 0016
December 09, 2014 11:53
What does it mean error 0016?
How can I remove the error alert?
In the manual written , but it doesn't help.
Please help me.
to rate and post a comment
Lenny Place
re: Philips IDI 5000 error 0016
December 10, 2014 08:41
0016 CDiagsMachine.c A board failed its bootup test. An error message describing which board
failed should precede this error. If not, run the PCB's Bootup Status util and
it will display which board failed. This is usually an actual board failure.
to display the second screen 1. Simultaneously press Superkey and F6 (the sixth key to the right in the top row of keys on the lower user interface) to display a second page of error information it should a message associated with the error banner telling you which board failed
to rate and post a comment
Gurgen
re: re: Philips IDI 5000 error 0016
December 10, 2014 09:00
Thank you a lot for reply. When I am Simultaneously pressing Superkey and F6 it is just disappear it doesn't shows second page.
to rate and post a comment
Gurgen
re: re: Philips IDI 5000 error 0016
December 11, 2014 10:04
Thank you for your anwer, when I am pressing Superkey and F6 it just makes error disappear, it doesn't show 2-nd page, as well I run test and it gives me an error .
What is the solution for this situation?
to rate and post a comment
Lenny Place
re: re: re: Philips IDI 5000 error 0016
December 12, 2014 08:25
if the system comes up press the setups key then select diagnostics, select show boot status report to show the boot status in this you should see the board causing the error.
to rate and post a comment
Gurgen
re: re: re: re: Philips IDI 5000 error 0016
December 12, 2014 10:09
Thank you for your anwer, I run test and it gives me an error .
What is the solution for this situation?
to rate and post a comment
Lenny Place
re: re: re: re: re: Philips IDI 5000 error 0016
December 12, 2014 11:02
what is the error?
to rate and post a comment
Gurgen
re: re: re: re: re: re: Philips IDI 5000 error 0016
December 12, 2014 12:17
Тhe error is .
to rate and post a comment
Lenny Place
re: re: re: re: re: re: re: Philips IDI 5000 error 0016
December 12, 2014 05:36
you can connect a service terminal to the CPU board on the rear compartment and using hyperterminal monitor the boot sequence.
to rate and post a comment
Gurgen
re: re: re: re: re: re: re: re: Philips IDI 5000 error 0016
December 14, 2014 10:37
Twice I typed the error name but it doesn't show in the message.
to rate and post a comment
Robert Celmer
re: Philips IDI 5000 error 0016
December 10, 2014 08:41
One of your channel boards is bad
to rate and post a comment
Rey Gonzalez
re: Philips IDI 5000 error 0016
December 10, 2014 08:42
You can press superkey + Frame rate twice it will go away. But most likely you will get it back because that error the is a pcb failing to boot.
For more help contact me rey@allultrasoundservices.com
to rate and post a comment