We have a problem Philips infinion 1.5 t MRI system.
Everything is working normally, but Some times it shows Error during scanning, "Unable calibrate RF Check anatomy at iso center RF status coil connection select coil modeā¦"...then image does not come, trying other coil no image (head,knee,large ,small coil )
Trying isa coil or body coil image good, never have a problem
Waiting for 1 or 2... hours and power off machine, machine is normally working.
we checked all of the trolley cables and receive mux no problem. I think have problem to SC1 SC2 channel, we could not resolve.
Any possible solutions are greatly appreciated.
Thanks
Selcuk Akin
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Selcuk Akin
Philips infinion
January 27, 2011 09:04
We have a problem Philips infinion 1.5 t MRI system.
Everything is working normally, but Some times it shows Error during scanning, "Unable calibrate RF Check anatomy at iso center RF status coil connection select coil modeā¦"...then image does not come, trying other coil no image (head,knee,large ,small coil )
Trying isa coil or body coil image good, never have a problem
Waiting for 1 or 2... hours and power off machine, machine is normally working.
we checked all of the trolley cables and receive mux no problem. I think have problem to SC1 SC2 channel, we could not resolve.
Any possible solutions are greatly appreciated.
Thanks
Selcuk Akin
to rate and post a comment
Sanchez Navarro
Re: Philips infinion
January 28, 2011 08:58
Did you check to +/- 48V Power Supply in PDU?
to rate and post a comment