Philips Intera 1.0 MRI

Tags: MRI

(107)
(6)
(31)

Milner A. Flecha

Philips Intera 1.0 MRI

December 26, 2013 05:06

I have a system with the following error message.

" high outflow water temperature detected by the coil heat exchanger"

Can anyody help me?

Thank you

Log inor Register

to rate and post a comment

(34)
(2)

Haluk Aktas

re: Philips Intera 1.0 MRI

December 27, 2013 01:26

Check your chiller and filters, motor couplings etc

Log inor Register

to rate and post a comment

(11)
(1)

David Doucet

re: Philips Intera 1.0 MRI

January 16, 2014 06:34

There's a filter in a canister inline with the gradient coil. you can put the canister in bypass to replace the filter. You could have a kink in the line. The strainer inside of the heat exchanger.

1 person liked this
Log inor Register

to rate and post a comment