Your Email:
Milner A. Flecha
Philips Intera 1.0 MRI
December 26, 2013 05:06
I have a system with the following error message. " high outflow water temperature detected by the coil heat exchanger" Can anyody help me? Thank you
to rate and post a comment
Haluk Aktas
re: Philips Intera 1.0 MRI
December 27, 2013 01:26
Check your chiller and filters, motor couplings etc
David Doucet
January 16, 2014 06:34
There's a filter in a canister inline with the gradient coil. you can put the canister in bypass to replace the filter. You could have a kink in the line. The strainer inside of the heat exchanger.
Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com
Milner A. Flecha
Philips Intera 1.0 MRI
December 26, 2013 05:06
I have a system with the following error message.
" high outflow water temperature detected by the coil heat exchanger"
Can anyody help me?
Thank you
to rate and post a comment
Haluk Aktas
re: Philips Intera 1.0 MRI
December 27, 2013 01:26
Check your chiller and filters, motor couplings etc
to rate and post a comment
David Doucet
re: Philips Intera 1.0 MRI
January 16, 2014 06:34
There's a filter in a canister inline with the gradient coil. you can put the canister in bypass to replace the filter. You could have a kink in the line. The strainer inside of the heat exchanger.
to rate and post a comment