Philips INtera 1.5T fail
September 05, 2017 04:26
When start the scan in a Philips Intera 1.5T MRI unit,
appear an error message: The patient Support is not connected. Please check power switch at MDU.
But the scan is made normally, only the patsupport don´t move automatically if is necessary.
Any idea to help?
Thank you
to rate and post a comment
Snoclimber
re: Philips INtera 1.5T fail
September 06, 2017 10:11
fiber cables. look behind the BEM box or misc. box. easy to get pulled out.
to rate and post a comment
Andrew Hodges
re: re: Philips INtera 1.5T fail
September 06, 2017 10:24
Can I check it only with light?
to rate and post a comment
Marshall Shannon
re: Philips INtera 1.5T fail
September 06, 2017 10:11
Bad fiber likely
to rate and post a comment
Andrew Hodges
re: re: Philips INtera 1.5T fail
September 06, 2017 10:24
Do you have for sale? Price? Thank you
to rate and post a comment
David Doucet
re: Philips INtera 1.5T fail
September 06, 2017 10:24
Check F6 in the MDU is on. Check the fuses and power supplies in the patient support are good. to calibrate the Support - press the manual table top release and move the table top manually in more the 100cm. Press the manual table top release again and now drive the table top fully out with the thumbwheel. Now drive the patient support down a little bit.
to rate and post a comment
Andrew Hodges
re: re: Philips INtera 1.5T fail
September 06, 2017 12:38
Thank you
to rate and post a comment