Fernando Gomez

Philips Intera Gradients Specs

December 09, 2011 08:36

I am looking for the tech specs for the gradient systems used in the Philips Intera MRI. Strength and slew rates mainly.
There are three of them. Omni, Power and Master.
fgomezvcl@yahoo.com

Log inor Register

to rate and post a comment

(1)

(4)

Heiko Kausch

Re: Philips Intera Gradients Specs

January 27, 2012 10:18

Hi here the tech specs:

gradient slew rate (T/m/s)
omni 20
power 75
master 150

gradient strength (mT/m)
omni 23
power 30
master 30