Mkrtich Altunyan
Philips iu22. error 511
April 22, 2019 09:35
Hi all. I have a problem on Philips iU22. Error code 511. I am looking for service password for recover data base.
Please help.
Hpservice1@gmail.com
to rate and post a comment
juan pablo amado
re: Philips iu22. error 511
April 24, 2019 10:16
511
Condition
Data Memory Manager Failed To Initialize.
Description
Platform Monitor failed to start the Data Memory Manager process.
Corrective Action
Possible cause: Local Study Repository (LSR) database corruption. From
the MTE run the DmmScript. NOTE: executing DmmScript erases all locally
stored patient study data
to rate and post a comment
juan pablo amado
re: Philips iu22. error 511
April 24, 2019 10:16
Try boot from Hirens boot CD, mini Wxp; then use windows tools to repair SO HDD