Philips Libra BV blank expose

Babar Beq

Philips Libra BV blank expose

December 12, 2016 10:25

Hello,

I have a Philips BV Libra which had problems with focusing so the exposed image was blurry. I opened the machine and found that the ribbon cable on the CCD was broken. I replaced the cable but now when I press expose I get a noisy blank image. Please provide me feeback on this issue or the service manual If you have it.

Regards,
Babar Beq

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Philips Libra BV blank expose

December 13, 2016 10:09

Check the harness at the camera, II area going to the Suco area this harness is know to have issues where the pins slip back on the connectors