Well, frustration with troubleshooting is not effective, only problem solving helps.
In principle, the height movement in all Philips C-arm systems has three switches. 2 limit switches and the collision switch. The collision switch is mounted in the back of the column, the limit switches in front of the column.
Purely from the error pattern, I would check the lower limit switch, I think it could be stuck.
The stand controller has a D-Sub connector with the connections. it can be measured well.
Comprehensive medical imaging support. We supply essential parts & maintenance for most OEMs, buy/sell quality used equipment, and offer mobile rentals globally. Contact us: info@lasoimaging.com
Gary Benitez
Philips Libra Stuck
April 18, 2022 11:33
Hi Guys,
I have a Philips Libra which the c-arm hydraulics that lifts and lowers the c-arm gets stuck six inches above the zero line mark.
I have tried tapping and holding down both switches a second time as required but it does not go lower.
It does go up but on the way down it goes no further the six inches from zero.
Fustrated.
Thanks,
to rate and post a comment
Herrmann
re: Philips Libra Stuck
April 19, 2022 09:24
Hi,
Well, frustration with troubleshooting is not effective, only problem solving helps.
In principle, the height movement in all Philips C-arm systems has three switches. 2 limit switches and the collision switch. The collision switch is mounted in the back of the column, the limit switches in front of the column.
Purely from the error pattern, I would check the lower limit switch, I think it could be stuck.
The stand controller has a D-Sub connector with the connections. it can be measured well.
I wish you success