Philips Medio 50 CP Error E335

October 26, 2009 10:47

Dear Colleague,
Have you ever had Error E335 with Philips Medio 50 CP -Xray generator ?! We have found several cases. Mostly problem disappear after warming time (about 5...25 minutes after switching ON the generator).
But also we have cases where problem doesn't go away (so, the E335 staying on). It seems that all old Medio 50 CP starting to have same problem...?!
Any idea ?!
Thanks,
Eero Huttunen, Service Technicial, Lifemed oy

Log inor Register

to rate and post a comment

JOSE ANTONIO SANCHEZ GALVAN

Re: Philips Medio 50 CP Error E335

October 29, 2009 01:41

dear colleague:
error 335 means "converter no ready" and i can guess there is a false contact somewhere in converter circuit. . .. check your equipment and, of course, your diagrams. . .