Roberto Villalba

Philips MRI 1.5T problem

May 26, 2009 09:49

New installation.
The images are displayed inverted; the scan protocol appear to be OK.

Any idea?

Thank you

Log inor Register

to rate and post a comment

(268)
(20)
(167)

Marshall Shannon

Re: Philips MRI 1.5T problem

May 26, 2009 10:22

YOUR GRADIENTS WERE HOOKED UP WRONG OR YOU HAVE + AND - REVERSED SOMEWHERE. DONT COUNT ON WIRE MARKERS, OHM OUT CABLES TO MAKE SURE MARKERS ARE RIGHT.

BUT NO DOUBT THAT IS WHERE YOUR PROBLEM IS.

INVERT USUALLY POINTS TO X SOMETIMES X AND Z

Log inor Register

to rate and post a comment

(57)

(1)

Bagher Mahdavi

Re: Philips MRI 1.5T problem

May 28, 2009 12:35

Hi all
Marshal is right.
there is no other fault.

Bagher