Philips MX8000 Quad scan stop problem

August 23, 2010 09:27

we have problem on this system .we are looking for service engineer for this problem.System is giving scan stop .we have also changed the xray tube.We have all spare parts.if somebody knows this system please copntact from following mail:bbtek@ada.net.tr

Log inor Register

to rate and post a comment

(373)
(66)
(15)

yuan bo

Re: Philips MX8000 Quad scan stop problem

August 23, 2010 07:11

i think you need to read the service manual, i have all , if you need any help, please contact obnauy@163.com

Log inor Register

to rate and post a comment

(57)
(1)
(15)

Barbaros Bulent Temiz

Re: Philips MX8000 Quad scan stop problem

August 24, 2010 08:51

we have sevice manual but we are looking service engineer to check that system.

Log inor Register

to rate and post a comment

(28)

Veronica Alvarez

Re: Philips MX8000 Quad scan stop problem

October 13, 2010 02:58

sytem is working ok now????? wew know the system and need the error log sent by email when you get the error then we can determine if it is the filament current or the h.v. or tube since you told in email that you changed the tube we can discard that and filament or h.v if the error log says that o.k send us email!!!!

Log inor Register

to rate and post a comment

saju bhaskaran

Re: Philips MX8000 Quad scan stop problem

September 19, 2010 11:56

is your system working now?

1 person liked this
Log inor Register

to rate and post a comment

stats
DOTmed text ad

Auction Alert - Live Now: Therapeutic Technologies Surplus Auctions June 15th

Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)