Over 500 Total Lots Up For Auction at One Location - NJ 06/15

philips practix 100 mobile x-ray

(7)
(3)

omer gafer

philips practix 100 mobile x-ray

August 28, 2020 10:27

please any one help me I have philips practix 100 after expossure the roter dos'nt stop still ranyng for long time

Log inor Register

to rate and post a comment

(243)
(18)
(9)

Carlos Duran

re: philips practix 100 mobile x-ray

August 31, 2020 12:23

the thyristor that energizes windings is in short circuit.
be carefull avoiding that condition, because you can burn the winding, becoming unusable whole monoblock.