Over 150 Missouri Auctions End Today - Bid Now

Philips Practix 3000 get stuck

Tags: Radiology - R/F, PHILIPS

(6)

Matias Canton

Philips Practix 3000 get stuck

October 05, 2013 09:07

Hello everybody, sorry about my english is not my natural language. I am having problems with a practix 300. When I pulse preparation the equip get stuck and nothing happen. If someone have been dealing with this equip knows that there is a signal call "KV enable" that if you look at the service manual says that appear when you power on the unit. I don´t have that signal and I suppose that is why when you try to prepares the equip doesn´t respond. I want to know to what parameters respond the availability of "kv enable" or if it is only softaware generated.

thanks a lot!!!
Biomedical Engineer Matias Canton

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Philips Practix 3000 get stuck

October 07, 2013 12:44

check your filaments

Log inor Register

to rate and post a comment

(6)

Matias Canton

re: re: Philips Practix 3000 get stuck

October 08, 2013 05:59

Thank for your time. The filaments are ok. You can see its trough the tube window and are lighting (of course one per time)

Log inor Register

to rate and post a comment

(6)

Matias Canton

re: Philips Practix 3000 get stuck

October 08, 2013 05:59

Thank for your time. The filaments are ok. You can see its trough the tube window and are lighting (of course one per time)

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your Trusted Multi-Vendor Imaging Service Partner

BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)