Can any one help me with the Optimus 50 generator,the problem in the AEC .It does not work on both the vertical Bucky and the horizontal table.At the vertical it gives low dose,at horizontal it gives very large dose.Any service manual will help.
Thanks,
Is it possible at the same time to loose both the values or the supplies in both vertical and horizontal-Are these conneted independent or together to the PCB-Basic Interface Z 150.Is it possible this PCB defected?
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
Marwan Kumsie
Philips Rad Bucky Diagnost problem
October 01, 2012 09:54
Can any one help me with the Optimus 50 generator,the problem in the AEC .It does not work on both the vertical Bucky and the horizontal table.At the vertical it gives low dose,at horizontal it gives very large dose.Any service manual will help.
to rate and post a comment
Lenny Place
re: Philips Rad Bucky Diagnost problem
October 01, 2012 09:09
is it possible you lost your Amplimat correction values in the adjustment setups? If they are good then check you Amplimat supplies
to rate and post a comment
Marwan Kumsie
re: re: Philips Rad Bucky Diagnost problem
October 03, 2012 09:48
Thanks,
Is it possible at the same time to loose both the values or the supplies in both vertical and horizontal-Are these conneted independent or together to the PCB-Basic Interface Z 150.Is it possible this PCB defected?