I am now getting this error 5 times out of ten on start up as the initial self test finishes, no probes are connected, ultrasound is not really my centre of excellence, but I'm open to suggestions.
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Malcolm Daw
Philips Sonos 5500 Error Code at start up EF8D4
February 15, 2016 07:32
I am now getting this error 5 times out of ten on start up as the initial self test finishes, no probes are connected, ultrasound is not really my centre of excellence, but I'm open to suggestions.
Many thanks. Malcolm.
to rate and post a comment
Alfred M. Brown
re: Philips Sonos 5500 Error Code at start up EF8D4
February 16, 2016 01:00
Run Time Error Code: E = Scanner. F = Non-recoverable. 8D4 = Front End Serial Interface test failed at Distribution Board.
to rate and post a comment
Malcolm Daw
re: re: Philips Sonos 5500 Error Code at start up EF8D4
February 17, 2016 03:44
Many thanks Alfred,I'll try to lay my hands on a dist.board and carry out a swap.