Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Philips Super 50CP error 3960

Tags: Radiology - R/F, PHILIPS

(28)
(1)

Ronald

Philips Super 50CP error 3960

October 16, 2013 04:40

Dear All.
Kindly help me to send the service manual for Philips Super 50CP generator.
Thanks and GBU all.
ronaldliem@yahoo.com

Log inor Register

to rate and post a comment

(62)
(20)
(8)

Muhammad Muhammad

re: Philips Super 50CP error 3960

October 18, 2013 10:14

Dear Contact me.

Log inor Register

to rate and post a comment

(17)
(6)

mab73

re: Philips Super 50CP error 3960

November 01, 2013 06:30

I got the manuals if u still need them

Log inor Register

to rate and post a comment

(62)
(20)
(8)

Muhammad Muhammad

re: re: Philips Super 50CP error 3960

November 01, 2013 10:14

EMERGENCY-OFF errors:
If hardware ready 2 (signal ready for exposure RD2) disappears after EXRQ was delivered or while the exposure is on
(EXON active), this is interpreted as EMERGENCY-OFF. The RD/EMOF signals of the following assemblies are tested
by the computer.

Log inor Register

to rate and post a comment

Amr Elsayed

re: re: Philips Super 50CP error 3960

January 12, 2018 03:19

Please , can u send me the service
Bme.amrelsayed@yahoo.com

Log inor Register

to rate and post a comment

DMEDS

re: re: re: Philips Super 50CP error 3960

January 17, 2018 10:35

Dear Amr,
i have the manuals for super cp generator ,
if you need it contact me for making offer for you,
magmesme@gmail.com

error 3264 :
Standby filament current, actual value small focus
from EH141 X1B:02 > ±35% tolerance measured
every second in standby and fluoro, during self test
after switch on

thanks

with best regards

Log inor Register

to rate and post a comment

Amr Elsayed

re: re: re: re: Philips Super 50CP error 3960

January 17, 2018 04:22

Dear Dmeds,

thank for you to offer your helping , i will be appreciated .
i contact with you from my mail ( bme.amrelsayed@yahoo.com) to receive the service

thanks again

Log inor Register

to rate and post a comment

Amr Elsayed

re: re: Philips Super 50CP error 3960

January 17, 2018 04:22

please i need the service manual
bme.amrelsayed@yahoo.com

Log inor Register