Philips Super 80CP Generator. Rotor problem

October 15, 2013 03:14

Hello: I have a problem with the rotor in a Super 80CP. The x-ray sequence looks ok and the x-rays are release but the rotor doesn´t run and the generator blocks with high kv or mA. I check with the undertable tube and the result is the same. I have made several checks and can´t find the problem. plese let me know if someone has an idea about this problem.

1 person liked this
Log inor Register

to rate and post a comment

(12)
(3)

jose antonio sanchez galvan

re: Philips Super 80CP Generator. Rotor problem

October 15, 2013 07:17

it is really extrange because if you have x ray, then all the secuence must be in order. . . check the rotor ready signal ,. . . this signal apears when rotor is in the righ speed. . do you hear rotor is running ?. . where are you writing from . .?
what error number do you get. ?
regards

jose a. sanchez g.

1 person liked this
Log inor Register

to rate and post a comment

(8)

Juan Manuel Esquivel Martinez

re: re: Philips Super 80CP Generator. Rotor problem

October 16, 2013 04:40

HI IM LOOKING PROGRAMMING MASK NUMBER 3.1 -3.2-3.3--3.4 FOR THIS MODEL X RAY SISTEM
COULD YOU HELP ME?
TKS

Log inor Register

to rate and post a comment

(8)

Juan Manuel Esquivel Martinez

re: re: Philips Super 80CP Generator. Rotor problem

October 16, 2013 04:41

HI I NEED PROGRAMING MASK NUMBER 3.1 UNTIL 3.5
DO YOU HAVE A COPY ABOUT IT
TKS
JUAN MANUEL

Log inor Register

to rate and post a comment

(8)

Juan Manuel Esquivel Martinez

re: Philips Super 80CP Generator. Rotor problem

October 16, 2013 04:41

YOU MUST CHECK USING THE PROGRAMMING BOOK LOOK ERROR NUMBER AND FIX IT USING PROPER PROGRAMING MASK

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)