tahar nouri

Philips Tomoscan Performance S x2a error

April 16, 2008 08:07

i am facing problem with philips tomoscan av performance s (AVPS) .i got x2a error even in warmup or air cal , i changed HT cables then i perform a test in stationary scanning i got the same problem;i noticed that there is leaking oil from HT tank cathode side. sometimes i if i try i get tube arcing error. is the problem in ht tank or tube.

Log inor Register

to rate and post a comment

saju bhaskaran

Re: Philips Tomoscan Performance S x2a error

April 23, 2008 03:46

You have to test HT tank without Xray tube first to confirm this.How to do the test is mentioned in the doccumentation
saju

1 person liked this
Log inor Register

to rate and post a comment

(158)
(30)
(1)

tahar nouri

Re: Philips Tomoscan Performance S x2a error

April 23, 2008 04:48

i appriciate well your help

you said that i can test HV tanks without tube
i did not find the procedure in documentation , please which part or section in technical guide is mentioned this test.

Log inor Register

to rate and post a comment

saju bhaskaran

Re: Philips Tomoscan Performance S x2a error

April 24, 2008 01:21

Dear Tahar
Long time I have not seen this machine .

Log inor Register

to rate and post a comment

saju bhaskaran

Re: Philips Tomoscan Performance S x2a error

April 24, 2008 05:55

you have to do dielectric test HT tank.Its there in the manual.
saju

Log inor Register

to rate and post a comment