Hello everyone,I have a Philips Trim III ECG machine having problem and the problem is that in place of leads number showing on screen it show "N" and no ECG wave is giving,if some one have solution please help me,
Regards,
Ghufran Ullah,
Pakistan
friend,,,,i changed all leads(limbs and long) but the error still maintain,,,In my hospital approx, there are 15 numbers of Philips Trim III ECG machine but some time we only wash the module PCb and heat it some of our recover by doing this not all,,,,,,
So friend is there any other possible solution instead of changing N lead if then please share.Thanks for your Cooporation
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
GHUFRAN Ullah
Philips Trim III ECG module
July 04, 2015 08:21
Hello everyone,I have a Philips Trim III ECG machine having problem and the problem is that in place of leads number showing on screen it show "N" and no ECG wave is giving,if some one have solution please help me,
Regards,
Ghufran Ullah,
Pakistan
to rate and post a comment
Firas Afifi
re: Philips Trim III ECG module
July 11, 2015 09:28
Hello, you need to change the N lead. It is cut.
to rate and post a comment
GHUFRAN Ullah
re: re: Philips Trim III ECG module
July 12, 2015 02:18
friend,,,,i changed all leads(limbs and long) but the error still maintain,,,In my hospital approx, there are 15 numbers of Philips Trim III ECG machine but some time we only wash the module PCb and heat it some of our recover by doing this not all,,,,,,
So friend is there any other possible solution instead of changing N lead if then please share.Thanks for your Cooporation