Would anyone have the tech lit for the Philips Veradius? We're getting a "M395 Battery OK, hardware error in generator" as soon as we try to expose.
The error is the same in the BV family manual (Pulsera, etc..) but the internal layout is clearly different.
Battery voltage is good. Filament glows at the tube. No voltage drop when we try to expose, but hear/feel relay engage. Anyone have insight or experience with this error?
The hospital had another Veradius we could take readings from. Swapped the entire ESU. It was good-to-go. Turned out it was the SEA5 board (not sure the exact Philips part number or name; we don't have a manual).
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Sang Un Choi
Philips Veradius manual? M395 error code
October 21, 2014 10:21
Would anyone have the tech lit for the Philips Veradius? We're getting a "M395 Battery OK, hardware error in generator" as soon as we try to expose.
The error is the same in the BV family manual (Pulsera, etc..) but the internal layout is clearly different.
Battery voltage is good. Filament glows at the tube. No voltage drop when we try to expose, but hear/feel relay engage. Anyone have insight or experience with this error?
Thanks
to rate and post a comment
Lenny Place
re: Philips Veradius manual? M395 error code
October 21, 2014 06:18
it is a ESU fault normally. I have found faulty diodes in the ESU. what is the part number of the ESU?
to rate and post a comment
Sang Un Choi
re: re: Philips Veradius manual? M395 error code
October 22, 2014 09:19
Thanks for the reply. It's what we were leaning towards. I'll have to get that info later today.
to rate and post a comment
Sang Un Choi
re: re: Philips Veradius manual? M395 error code
October 23, 2014 03:34
The hospital had another Veradius we could take readings from. Swapped the entire ESU. It was good-to-go. Turned out it was the SEA5 board (not sure the exact Philips part number or name; we don't have a manual).
to rate and post a comment
Lenny Place
re: re: re: Philips Veradius manual? M395 error code
October 23, 2014 04:18
I can send you the manual for the generator. would need an email address
to rate and post a comment
Sang Un Choi
re: re: re: re: Philips Veradius manual? M395 error code
October 23, 2014 10:49
Sang.ezekiel.choi@gmail.com
Would that have part numbers for boards in the card cage?
to rate and post a comment
Sang Un Choi
re: re: re: re: Philips Veradius manual? M395 error code
November 03, 2014 08:26
sang.ezekiel.choi@gmail.com
to rate and post a comment