Philips Veradius xray lamp error

Tags: C-Arm, PHILIPS

(9)
(1)

Chris Flores

Philips Veradius xray lamp error

April 04, 2018 04:22

Intermittently receiving "X ray lamp error, system still available"

Log inor Register

to rate and post a comment

(49)
(13)

Brad

re: Philips Veradius xray lamp error

April 05, 2018 03:00

Possible tube arcing? Have you checked high voltage cables for indication of arcing? Not sure what lamp other than charging indicator is on the system.

Log inor Register

to rate and post a comment

(82)
(15)

Herrmann

re: Philips Veradius xray lamp error

April 06, 2018 10:01

Dear Chris,

Dear Chris,

Please check the lamp socket and the lamp itself. This can lead to this problem. The lamp is controlled by the USB BOB. There too, a plug can be loose. (WBOB X9 or X16)

Many greetings
Willi
-------

Incidentally, if there is a problem with a Philips BV system, please always include the 6NC number.

This starts with Philips BV system always with 718 ??? and only with this number is a device technically identifiable and it can be worked with the proper documentation.

The software version is also important!

Log inor Register

to rate and post a comment

stats
DOTmed text ad

JOHNSON & JOHNSON MONARCH QUEST - For Sale $299k or Lease $6k Per Month, 48 Months

Lightly used system available direct from the end user customer. MONARCH QUEST Featuring our most advanced, data-driven navigation software with AI-Powered algorithms & advanced imaging integration with the GE HealthCare OEC 3D C-arm.

stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)