Dear all

I have a problem at philips briliance 6 slice ct .when machine on this message appear (Gantry acqisition sys can not comply .please retry)
After awhile at warm up this message appear (scan 14 out of 27).

Thanks for any help
Best regards
Sara.ahmed.hazem@gmail.com

Log inor Register

to rate and post a comment

Dustin Zimmerman

re: Philips-Brilians

August 04, 2016 11:01

You have a data transfer or receive problem . Could be your aqquisitor board in the IRS or faulty transmitter, waveguide, or prism pickup. Also could be rcom. Check in the s1 irs server event log for errors. Run diagnostics on acquisitor board via rdp into s1.

Log inor Register

to rate and post a comment

(29)
(4)
(4)

William Dixon

re: Philips-Brilians

August 04, 2016 11:01

Sara

after the warmup fails or completes
You need to open up the ERROR LOG VIEWER from the left hand side menu.
read the most right hand column -- the messages column

in there you may have some error codes that you can reply back to this post with so we can tell what may be failing

Log inor Register

to rate and post a comment

(138)
(18)
(1)

umberto sciunnach

re: Philips-Brilians

August 04, 2016 11:23

Hello, you have a rotation problem, ypu can see also durin warm up.
The posotion os not correct.
Test of rotation.
Regards
Umberto

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Philips-Brilians

August 04, 2016 11:29

i would first start by inspected the slip rings and brushes. what is the actual error message in the error log?

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT PARTS USED-TESTED, REFURBISHED. Call (512) 477-1500

Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com

DOTmed Certified stats

Upcoming Auction Events

Therapeutic Technologies
TIMED
Jun 15, 2026 (538 lots)