Dear colleagues
I have a problem in Phillips ct 16 slice.
My problem is, sometimes after the 3rd or 4 th patient the Guntry stopped suddenly, and this messages appears:
S-GEN-FAULT-ANALOG-FAULT
S-RHOST-HVG-GENERATOR-ARC-FAULT
S-GEN-FAULT-CATHODE-TUBE-ARC
S-GEN-FAULT-ANODE-TUBE-ARC
Make sure that your Visa Key is active...
Perform the NO LOAD test to determine the ARC origin ... whether tube or generator The results of the test are encoded in the GANTRY_INFO_ERROR_BKST and RHOST_NORMAL_ERROR_BKST messages. The strings associated with each message provide information regarding the results of the test. If there is an arc, you will see APM Arc or CPM Arc in the string. Both the Anode and Cathode Power Modules will detect an arc in the tube so you will see both APM Arc and CPM Arc strings in the messages.
From your error messages, you have both APM and CPM arc strings which means most likely the arcing is coming from the tube
You can contact me for manuals and further assistance
Therapeutic Technologies closeout auctions take place on DOTmed.com. Bid today on 538 lots of medical equipment. Inspections may be done by appointment. Hours are Monday through Friday 9am-4pm Contact chrisnuytens1@gmail.com to schedule.
a.h.m
Phillips ct 16- slice stopped
September 28, 2023 06:51
Dear colleagues
I have a problem in Phillips ct 16 slice.
My problem is, sometimes after the 3rd or 4 th patient the Guntry stopped suddenly, and this messages appears:
S-GEN-FAULT-ANALOG-FAULT
S-RHOST-HVG-GENERATOR-ARC-FAULT
S-GEN-FAULT-CATHODE-TUBE-ARC
S-GEN-FAULT-ANODE-TUBE-ARC
Any one can help me, please?
Best regards
to rate and post a comment
CRAIG WAMBURA
re: Phillips ct 16- slice stopped
October 02, 2023 09:26
Make sure that your Visa Key is active...
Perform the NO LOAD test to determine the ARC origin ... whether tube or generator The results of the test are encoded in the GANTRY_INFO_ERROR_BKST and RHOST_NORMAL_ERROR_BKST messages. The strings associated with each message provide information regarding the results of the test. If there is an arc, you will see APM Arc or CPM Arc in the string. Both the Anode and Cathode Power Modules will detect an arc in the tube so you will see both APM Arc and CPM Arc strings in the messages.
From your error messages, you have both APM and CPM arc strings which means most likely the arcing is coming from the tube
You can contact me for manuals and further assistance
to rate and post a comment