Phillips ct 16- slice table problem

Tags: CT

(21)
(5)

a.h.m

Phillips ct 16- slice table problem

August 17, 2023 04:24

Dear colleagues
I have a problem in Phillips ct table vertical movement up & down.
When power on the system and push up or down bottom the table move about 1 cm and stopped.
And this message appears in error log viewer:
(S-MC-Device-Couch-Vertical-Hw-Fault)
Sometimes,After making restart, the table didn't move and this message appears
(S-Mc-Device-PS-Failure).

Thanks for any help.
Best regards.

Log inor Register

to rate and post a comment

(105)
(17)
(17)

Doscar Ernesto Perez Pereyra

re: Phillips ct 16- slice table problem

August 18, 2023 11:41

Easy one...

Replace Vertical Inverter...

1 person liked this
Log inor Register

to rate and post a comment

(21)
(5)

a.h.m

re: re: Phillips ct 16- slice table problem

August 18, 2023 02:40

Thanks for replying Mr. Doscar
In service mode the table movement
Is good ,up and down so I think inverter board is ok.

Log inor Register

to rate and post a comment

(105)
(17)
(17)

Doscar Ernesto Perez Pereyra

re: re: re: Phillips ct 16- slice table problem

August 21, 2023 11:48

That is the problem for people who do not take the training and do not know the theory... the service mode directly applies the 220V to the K101 relay that activates the frequency of descent and ascent... but it does not act on the logic, which manages the canbus to go down and go up. (In theory, even if it goes up and down in service mode it doesn't mean that it works in normal mode)

Log inor Register

to rate and post a comment

(12)
(2)

Nikesh Chitti

re:re: re: Phillips ct 16- slice table problem

August 21, 2023 11:50

when Doscar says , just follow it. in manual mode system don't use vertical inverter for table movement.

1 person liked this
Log inor Register

to rate and post a comment

(21)
(5)

a.h.m

re: re: Phillips ct 16- slice table problem

September 28, 2023 10:55

Thank you very much Mr.Doscar.
I replaced the inverter, and it's ok.
Your advice is perfect.
Thank you again.

Log inor Register

to rate and post a comment

CMSService

re: Phillips ct 16- slice table problem

August 18, 2023 02:32

Hello, should see the error description, anyway you can check power supply on couch.

Log inor Register

to rate and post a comment

(21)
(5)

a.h.m

re: re: Phillips ct 16- slice table problem

August 18, 2023 02:41

Thanks for replying
I checked, its ok.

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT PARTS USED-TESTED, REFURBISHED. Call (512) 477-1500

Extensive inventory of Philips & Siemens MRI & CT used-tested, refurbished parts. Knowledgeable Sales and Inventory Team. 90-Day Warranty. Bilingual Support. Competitive Pricing. Technical Support. ISO Certified. parts@keimedicalimaging.com

DOTmed Certified stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)