Over 150 Missouri Auctions End Today - Bid Now

Picker 5000 problem

(17)

Mat Isa Zulkhairi

Picker 5000 problem

January 21, 2010 10:22

Contactor Off : TURN KEYSWICTH TO 'SYST' AND TRY AGAIN

I need some help, I dealing with Picker 5000, I having a problem with Contactor.I could'nt warming the system because Contactor Off and also the ready light on the Console does not lit.

please advise

Log inor Register

to rate and post a comment

(2)

Alonzo Wilson

Re: Picker 5000 problem

January 21, 2010 01:44

Check the power suppys in back of Gantry 5v 15 -15
with a scope look for ripple.
Al Wilson

Log inor Register

to rate and post a comment

(17)

Mat Isa Zulkhairi

Re: Picker 5000 problem

February 12, 2010 09:26

Hai Alonzo,

Problem fixed after I changed 1 minifuse on the XSC Board, Now 'Ready" light Lit.

Thank you Alonzo

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

Re: Picker 5000 problem

January 21, 2010 10:11

hi Matt
there are a lot of things that can avoid you have green light. there will be no operation until green light turn on.
Open front door of gantry, set BOTH secure lines, pull the gullwing sw until you listen a click and try again. on AC INUT BLOCK you will see some leds green and yellows. let me know which are lit on this module and on system board module. make sure of the led lit name, some errors can ocur seing it from an angle.
check ac input voltaje. it must be 450 or more.
check there are four leds lit on back of gantry at left of the service normal select key.
check all fuses you can see WARNING: TURN OFF EQUIPMENT FROM THE WALL BEFORE ATTEMP IT.
Feel free to contact me at the email for further help.
good luck
jose

Log inor Register

to rate and post a comment

(17)

Mat Isa Zulkhairi

Re: Picker 5000 problem

January 22, 2010 09:40

These are the list what I have done today:

1- I changed Signal slip brushes and Brushes, Then clean up the slip ring.
2- Checked all the fuses at back of the gantry, all look good.
3- I found 2 relays Lit on the AC Power Supply (See Image 1,2 & 3) Is That Relay ?
4- Changed Ribbon Cable and reseated the SIF Board but not a Ribbon cable and SIF cause a problem,Problem still remain the same.
5- I found 2 AC Contactor on the AC Power Supply Activated when I turn The Key swicth.
6- There is no Fault LED light is Lit and all XSC power were present. If the XSC got power meaning the fuse in the XSC assy not blown? right ?
7- Gantry Rotate.

Problem Still remain the same.

Regard

Zulkhairi

Log inor Register

to rate and post a comment

(10)
(1)

mingyou fu

Re: Picker 5000 problem

January 23, 2010 01:07

1.check the fuse on XSC+ Board
2.check the AC Diagnostic board of the AC input module

Log inor Register

to rate and post a comment

(17)

Mat Isa Zulkhairi

Re: Picker 5000 problem

February 08, 2010 09:23

Hai Jose,

Problem fixed after I changed 1 minifuse on the XSC Board,

Thank you jose

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

Re: Picker 5000 problem

February 08, 2010 06:44

IT WAS PLEASURE TO HELP YOU

JOSE

Log inor Register