Picker Clinix E Table

Tags: Radiology - R/F, PICKER

(140)
(35)

FranciscoG

Picker Clinix E Table

August 07, 2013 03:43

Hi there,

I have an issue with a Picker Clinix E table. The table top magnetic locks don't get activated when the table is on, I checked underneath the table and saw a 2Amp 250 Volts fuse that was blown up so I changed it 2 times and keeps blowing it. Any help on how to get this fixed, is appreciated.

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Picker Clinix E Table

August 07, 2013 05:54

check the diodes across the magnetic locks it could be shorted or you have a bad lock

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: Picker Clinix E Table

September 27, 2013 07:01

Thank you Lenny! I changed the Transformer on the buttom of the table and the wiring in the box on the side of the table that were having a short. After all this, the table locks are great!

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: re: re: Picker Clinix E Table

September 28, 2013 09:16

thank you for the update. I will keep your information

Log inor Register

to rate and post a comment

(97)
(14)

Donald Bradbury

re: Picker Clinix E Table

August 09, 2013 08:16

Check the diodes or the lock coil is bad. These run pretty hot. I have seen people locate a fan next to them to help. As I remember these are a pain to replace involving dropping the tube stand (shoot the design engineer) Have help. Good luck

1 person liked this
Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: re: Picker Clinix E Table

September 27, 2013 07:01

Thanks Donald, you can read in the comment above what I did to have the locks working again.

Log inor Register

to rate and post a comment

(140)
(35)

FranciscoG

re: Picker Clinix E Table

September 27, 2013 07:01

Just in case someone needs some help with this... there was a short in the box on the side of the table that connects to the table release button, the 120-24V transformer on the buttom of the table was replaced as well because this short was blowing the fuses up!
Finally the locks are working great!

Log inor Register

to rate and post a comment

stats
DOTmed text ad

Your one-stop source for all Pre-Owned Medical Imaging Equipment needs. 📞 305-234-0084

We supply high-quality refurbished and used imaging equipment and parts to hospitals, imaging centers, and brokers worldwide. We pay top dollar and offer installation, deinstallation, service, and transport. Se habla español info@medilabglobal.com

DOTmed Certified DOTmed 100 stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)