Hi, we are having problems with a pq2000(mark III tube)when we tried to perform any study this message show up "error 1017, bad exchanger/starter/cables/oil lines-no oil flow" we d performed the seasoning and the pre-heat with no problems, we also checked the rings and power brush(ring test) any help would be apreciate
Hi: well is like the gentlemen already advice you; your problem is on the high speed started. in the cases I have seen 99% of errors 1017 have been bad High speed starter. Let me know if you need more help.
Your original problem it is actually air inside the x ray tube, you should refill the x ray tube
before making any exposure.
If you try expose bypassing the flow switch you are gonna damage the inverters modules, cause the oil function is both, cooling and isolation.
I would better short it just for testing purposes. If the bypass makes it work, then replace the flow sensor. If you keep the sensor bypassed, in the future you may have a real oil flow problem and destroy the XRay tube without notice it until it's too late.
we had shorted the flow sensor, and managed to perform a study, but it burned the inverter driver control :( , is it possible to fix if there is not oil flow?
Blanca Parciak
Picker ct PQ error 1017
May 03, 2006 12:42
Hi, we are having problems with a pq2000(mark III tube)when we tried to perform any study this message show up "error 1017, bad exchanger/starter/cables/oil lines-no oil flow" we d performed the seasoning and the pre-heat with no problems, we also checked the rings and power brush(ring test) any help would be apreciate
to rate and post a comment
Marshall Shannon
Re: Picker ct PQ error 1017
May 03, 2006 01:20
Error refers to the XSC (Generator)
High_Speed Starter_Driver
Possible bad XSC, DIB, HVG, Starter, exchanger, cables, fuses, estop pressed, low line, slip rings.
to rate and post a comment
Ernesto Guerra
Re: Picker ct PQ error 1017
May 03, 2006 03:27
Hi: well is like the gentlemen already advice you; your problem is on the high speed started. in the cases I have seen 99% of errors 1017 have been bad High speed starter. Let me know if you need more help.
to rate and post a comment
Blanca Parciak
Re: Picker ct PQ error 1017
May 03, 2006 06:33
thank you so much for your help, what does it mean? do i have to replace the whole high speed starter??
to rate and post a comment
Denis Alves
Re: Picker ct PQ error 1017
October 05, 2006 02:04
Your original problem it is actually air inside the x ray tube, you should refill the x ray tube
before making any exposure.
If you try expose bypassing the flow switch you are gonna damage the inverters modules, cause the oil function is both, cooling and isolation.
garuda66@gmail.com
to rate and post a comment
Iftikhar Khan
Re: Picker ct PQ error 1017
May 04, 2006 01:07
dear
before replacing the high speed starter, bypass (short) the flow sensor in heat exchanger, this might solve ur problem. this sensor often fails.
good luck
khan
to rate and post a comment
Sergio Leonardo Barral
Re: Picker ct PQ error 1017
May 04, 2006 08:29
I would better short it just for testing purposes. If the bypass makes it work, then replace the flow sensor. If you keep the sensor bypassed, in the future you may have a real oil flow problem and destroy the XRay tube without notice it until it's too late.
Best regards
Sergio
to rate and post a comment
Blanca Parciak
Re: Picker ct PQ error 1017
May 04, 2006 11:38
Hi
we had shorted the flow sensor, and managed to perform a study, but it burned the inverter driver control :( , is it possible to fix if there is not oil flow?
thks
to rate and post a comment