Over 150 Missouri Auctions End Today - Bid Now

Picker Eclipse 0xa T/R status error

Kadir Kuran

Picker Eclipse 0xa T/R status error

September 28, 2011 09:03


we have problem on Picker Eclipse mrii when we use body coil system give 0xa tr status eror we check body coil and we change T/R interface ,rxmux ,t/r switch .
Is any body have suggestion

Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

Re: Picker Eclipse 0xa T/R status error

September 28, 2011 05:30

Have you checked the Body coil diodes, by disconnecting J657 and ohming out between the center conductor to shield, it should read 57K if it reads 54K then possible shorted diode

Log inor Register

to rate and post a comment

Kadir Kuran

Re: Picker Eclipse 0xa T/R status error

September 29, 2011 08:39

Lenny

it is 57K looks all diots are okay

Log inor Register

to rate and post a comment

(21)
(3)

Pi Jaworski

Re: Picker Eclipse 0xa T/R status error

October 05, 2011 02:04

Did you check the diodes (i. e. 57 kohm) directly from the T/R switch? It happened to me twice that the BNC plug caused the intermittent problem.