If you have a broken Sonosite, Mindray, GE, Philips, Hitachi, or Aloka system or probe contact us today. Free ultrasound & probe EVALS with 1 Year Warranty on all repairs. We repair all orders in 1-5 days. Call 844-644-5440 Email: Sales@Lenard.Healthcare
Kadir Kuran
Picker Eclipse 0xa T/R status error
September 28, 2011 09:03
we have problem on Picker Eclipse mrii when we use body coil system give 0xa tr status eror we check body coil and we change T/R interface ,rxmux ,t/r switch .
Is any body have suggestion
to rate and post a comment
Lenny Place
Re: Picker Eclipse 0xa T/R status error
September 28, 2011 05:30
Have you checked the Body coil diodes, by disconnecting J657 and ohming out between the center conductor to shield, it should read 57K if it reads 54K then possible shorted diode
to rate and post a comment
Kadir Kuran
Re: Picker Eclipse 0xa T/R status error
September 29, 2011 08:39
Lenny
it is 57K looks all diots are okay
to rate and post a comment
Pi Jaworski
Re: Picker Eclipse 0xa T/R status error
October 05, 2011 02:04
Did you check the diodes (i. e. 57 kohm) directly from the T/R switch? It happened to me twice that the BNC plug caused the intermittent problem.
to rate and post a comment