DEar Friends
if someone has had the problem with error code E95 with a portable x ray machine Picker explorere II and want to share with me the troubleshooting to fix this faillure any help will be very appreciatted.
annorys1971@att.net.mx
that error mean the kv is low, usually that coming from battery bad, conect at gnd and (b7+) meter recorder max/min ie fluke and set the meter for min. then make exposure if the voltaje get down at least 15 percent change the bat
Hipolito Jorge Iglesias
picker explorer II error E95
March 01, 2012 09:58
DEar Friends
if someone has had the problem with error code E95 with a portable x ray machine Picker explorere II and want to share with me the troubleshooting to fix this faillure any help will be very appreciatted.
annorys1971@att.net.mx
to rate and post a comment
Joel Grullon
Re: picker explorer II error E95
March 07, 2012 06:04
that error mean the kv is low, usually that coming from battery bad, conect at gnd and (b7+) meter recorder max/min ie fluke and set the meter for min. then make exposure if the voltaje get down at least 15 percent change the bat
best regards
joel
to rate and post a comment
Hipolito Jorge Iglesias
Re: picker explorer II error E95
March 07, 2012 10:32
Dear Joel
to rate and post a comment