Have a couple of Picker IQ's in Mexico with expired diagnostics. Does anyone know if the monitor board chip is still availabe for purchase to unlock the diagnostics? thank you Herman Rice
If the chip in the machine says it has diagnostics, but they have expired, you can unlock them. Phillips won't sell a chip with diagnostics enabled unless they do a service contract on the machine. You can call me if you want.
Jeff
402-960-7231
You can enable the expired diags if you have the battery style system monitor PCB. It will have a round coin sized battery right next to the microcontroller chip. From IQ>show_cnf , get the expiration date of the chip. Then look in IQ>cal_del , select F6 just to view cal files, not delete, and get the most recent date.
Set the date before that date , but after the latest CAL file with the touch panel, then power down. short the + to - on the battery for at least 10 seconds. On power up green screen menu, select option 6 (warning will delete any images so be sure they are archived already) , and when you boot up diagnostics will work. If they dont repeat this procedure. If you have the NVRAM style board, Philips will sell you an "O" level chip for image cals only, for a cost of ~$250 USD
Good luck, Dave
I am a certified Picker ct field service engineer . I can help you to enable the diagnostic s/w . I do it to my systems whenever I have to calibrate them . If you still need my help please contact me . My e-mail is : abdusaed@link.net , mobile phone :+20122889193
I deed several times the procedure, but still diagnostics software is dissabled. I think there is some hardware throuble on System Monitor board.
My main problem on the PQS scanner is that CCT/SS Hardware is dissabled, ( the Gantry do not enable any drive ). Nothing seems to be
wrong on Gantry. But Console do not read XSC or GPU. I can do gxd comand to download software on XSC/GPU with 100% sucsesfull. I did change GPU, PMDC and Scanner Interface bds, with no throuble change. Do You have any idea what is happenig?.
I checked the interconenting cables also, and I can see some signal on reception/transmition optical serial communications lines on GPU.
You didn't need CCT/SS hardware to be enabled. This is an option.
The procedure to enable diag is correct and working. Look again for calibration data.
How you check that the console didn't read XSC and GPU.
Describe better your problem and the what checks are done
When I type show_cnf, at the botton of first page
appears the message: XSC can“t be read. ( slip ring/ brushes,fuses, etc, are ok ). The ready LED on OP Panel console gets on.
When I try to do a warm up, on the touch screen the message:
"Interrupted prosses, Retry 3".
On Regulus sceen: "Not time to transfer"
Some times: GPU corrupt code, download required
Setting the date in REGULUS will NOT work, you MUST set the date at the touch screen to your right.
The procedure will NOT work if your setting the date with the DATE command within REGULUS.
Try the procedure above again, this time short the battery prior to starting the procedure and leave it shorted until after the procedure, power-down for 15 minutes then power up again after the reset and date changes.
Looking for the leader in urology? BK Medical is and we are the 3rd party experts. We sell BK and many other brands. Call today to discuss your urology equipment needs and save $1,000’s! Call Today: 1-888-838-7488 Email: sales@sentinelimaging.net
Herman Rice
Picker IQ CT Scanner Diagnostic Software
April 13, 2005 10:15
Have a couple of Picker IQ's in Mexico with expired diagnostics. Does anyone know if the monitor board chip is still availabe for purchase to unlock the diagnostics? thank you Herman Rice
to rate and post a comment
Jeff Stuttle
Re: Picker IQ CT Scanner Diagnostic Software
April 14, 2005 09:12
If the chip in the machine says it has diagnostics, but they have expired, you can unlock them. Phillips won't sell a chip with diagnostics enabled unless they do a service contract on the machine. You can call me if you want.
Jeff
402-960-7231
to rate and post a comment
Dave Adams
Re: Picker IQ CT Scanner Diagnostic Software
April 14, 2005 02:46
You can enable the expired diags if you have the battery style system monitor PCB. It will have a round coin sized battery right next to the microcontroller chip. From IQ>show_cnf , get the expiration date of the chip. Then look in IQ>cal_del , select F6 just to view cal files, not delete, and get the most recent date.
Set the date before that date , but after the latest CAL file with the touch panel, then power down. short the + to - on the battery for at least 10 seconds. On power up green screen menu, select option 6 (warning will delete any images so be sure they are archived already) , and when you boot up diagnostics will work. If they dont repeat this procedure. If you have the NVRAM style board, Philips will sell you an "O" level chip for image cals only, for a cost of ~$250 USD
Good luck, Dave
to rate and post a comment
gustavo fiochi
Re: Picker IQ CT Scanner Diagnostic Software
April 15, 2005 07:57
you can enable the diagnostics. give me your phone, i'll be in mexico next week
i speek spanish
to rate and post a comment
Abdula Saeed El Shafeey
Re: Picker IQ CT Scanner Diagnostic Software
July 13, 2005 11:56
I am a certified Picker ct field service engineer . I can help you to enable the diagnostic s/w . I do it to my systems whenever I have to calibrate them . If you still need my help please contact me . My e-mail is : abdusaed@link.net , mobile phone :+20122889193
to rate and post a comment
Miguel Angel Lemus
Re: Picker IQ CT Scanner Diagnostic Software
February 26, 2009 06:00
Dear Abdula,
I need to enable the diagnostic software on a PQS CT System.
I have performed the procedure refered on this email forum. But I have had not good luck.
Do you have another tip on the procedure than you may share with me please ?
Many thanks,
Regards,
Miguel A. Lemus
to rate and post a comment
Ralph Childs
Re: Picker IQ CT Scanner Diagnostic Software
February 27, 2009 01:34
Miguel,
Please forward the procedure at the top of this page. Dave is absolutely correct about how to enable the software for service mode.
Scooter
scanner1@eim.ae
to rate and post a comment
Miguel Angel Lemus
Re: Picker IQ CT Scanner Diagnostic Software
February 28, 2009 11:59
Dear Ralph/
Scooter
I deed several times the procedure, but still diagnostics software is dissabled. I think there is some hardware throuble on System Monitor board.
My main problem on the PQS scanner is that CCT/SS Hardware is dissabled, ( the Gantry do not enable any drive ). Nothing seems to be
wrong on Gantry. But Console do not read XSC or GPU. I can do gxd comand to download software on XSC/GPU with 100% sucsesfull. I did change GPU, PMDC and Scanner Interface bds, with no throuble change. Do You have any idea what is happenig?.
I checked the interconenting cables also, and I can see some signal on reception/transmition optical serial communications lines on GPU.
Thank you very much for some body else comments,
Regards,
Miguel Lemus
biomel@navegante.com.sv
to rate and post a comment
Alexander Tonev
Re: Picker IQ CT Scanner Diagnostic Software
March 01, 2009 04:18
You didn't need CCT/SS hardware to be enabled. This is an option.
The procedure to enable diag is correct and working. Look again for calibration data.
How you check that the console didn't read XSC and GPU.
Describe better your problem and the what checks are done
to rate and post a comment
Miguel Angel Lemus
Re: Picker IQ CT Scanner Diagnostic Software
March 01, 2009 11:53
When I type show_cnf, at the botton of first page
appears the message: XSC can“t be read. ( slip ring/ brushes,fuses, etc, are ok ). The ready LED on OP Panel console gets on.
When I try to do a warm up, on the touch screen the message:
"Interrupted prosses, Retry 3".
On Regulus sceen: "Not time to transfer"
Some times: GPU corrupt code, download required
to rate and post a comment
Marshall Shannon
Re: Picker IQ CT Scanner Diagnostic Software
March 01, 2009 06:03
Setting the date in REGULUS will NOT work, you MUST set the date at the touch screen to your right.
The procedure will NOT work if your setting the date with the DATE command within REGULUS.
Try the procedure above again, this time short the battery prior to starting the procedure and leave it shorted until after the procedure, power-down for 15 minutes then power up again after the reset and date changes.
to rate and post a comment