kV balance errors normally are born in defective multipliers, more scarcely in cabling or analog board. The problem is to determine which multiplier is the faulty one. If you have the Picker Troubleshooting manual, follow the instructions regarding the cabling. But if you reach the point to measure the feedback you certainly can be unsure which one is the faulty multiplier because the HV generator shuts down before you can get a significative measurement even using a scope. At that point I suggest you to try the multipliers in another scanner or have spare ones and alternativelly replace them until you fix the problem.
HIGH QUALITY fully tested & refurbished systems to manufacturer specifications like new. Installation & service support worldwide. Parts & tubes in stock with warranty. Sales@Elsmed.com
Jane Hall
Picker PQ 1012 Error
April 17, 2007 12:36
HV Generator KV Balance Error ? Anode & Cathode KV diff ? How can we fix it?
to rate and post a comment
Edward Yan
Re: Picker PQ 1012 Error
April 17, 2007 02:25
Most likely you need a new x-ray tube replacement.
Edward Yan, B.S.E.E.
to rate and post a comment
gustavo fiochi
Re: Picker PQ 1012 Error
April 17, 2007 04:09
look at the erp or erl. if you are having ANA, most probably it is the tube. how many shoots does it have?
to rate and post a comment
Sergio Leonardo Barral
Re: Picker PQ 1012 Error
April 18, 2007 11:33
kV balance errors normally are born in defective multipliers, more scarcely in cabling or analog board. The problem is to determine which multiplier is the faulty one. If you have the Picker Troubleshooting manual, follow the instructions regarding the cabling. But if you reach the point to measure the feedback you certainly can be unsure which one is the faulty multiplier because the HV generator shuts down before you can get a significative measurement even using a scope. At that point I suggest you to try the multipliers in another scanner or have spare ones and alternativelly replace them until you fix the problem.
to rate and post a comment
Jim Foros
Re: Picker PQ 1012 Error
May 03, 2007 03:32
I have had this several times, it was always a multiplier.
to rate and post a comment