Over 500 Total Lots Up For Auction at One Location - NJ 06/15

Picker PQ 2000, Artifacts problems, Bad Detectors

Tags: CT

Rainner Lopez

Picker PQ 2000, Artifacts problems, Bad Detectors

May 01, 2018 01:29

Good day everyone, i'm having trouble with some bad detectors, i want to know which detectors are bad to can replace it, if anyone can send me the service manuals, and how to enter in service mode..

Best regards
rainner.lopez@hotmail.com

Log inor Register

to rate and post a comment

(6)

Alfredo Fuhr

re: Picker PQ 2000, Artifacts problems, Bad Detectors

May 02, 2018 09:54

Hello Rainner:
It is normally needed to have diagnostics enabled. Login as root, no password. Check this with the "rev" command. The bad channels can be visualized with the "dita" program. And they can be badlisted with "bad_dets" or the bad module(s) replaced outright. For details email me to alfredofuhr@gmail.com. Good luck