May 20, 2018 03:17

hi everyone, i hope anyone can help me with that

601 - Stop hit while initializing for a scan. (operator changed her mind)
Refer to the ERP user's manual in the Diag manual for further details.

the CT throws this error when i try to do a pilot or a scan.

and sometimes the system throws the error 512.

thanks in advance.

rainner Lopez, Venezuela
email: rainner.lopez@hotmail.com

Log inor Register

to rate and post a comment

(11)
(1)

Alfred65

re: picker pq 2000 Error 601

May 21, 2018 09:34

Hello Rainner
Error code 601 means that the recon boards in the console are stopped. Check the LEDs on the boards. Remove additional AP or EB boards. Minimum configuration for an axial scan is SIF, one AP, one BP.
Good luck

Log inor Register

to rate and post a comment

(13)

(12)

Jody Rotenizer

re: picker pq 2000 Error 601

May 21, 2018 09:36

Have you cleaned the slip ring?

Log inor Register

to rate and post a comment

(7)

rockbob

re: picker pq 2000 Error 601

May 21, 2018 05:59

clean the slip rings and brushes

Log inor Register

to rate and post a comment

(183)
(13)
(2)

Borhan Kalash

re: picker pq 2000 Error 601

June 05, 2018 09:43

try reconstructing good rawdata
is image ok?

Log inor Register