Over 150 Missouri Auctions End Today - Bid Now

Picker PQ200 CT system problem

Mark Adler

Picker PQ200 CT system problem

April 26, 2010 09:52

I have a problem in Picker PQ200 CT system . All Plot scans have vertical line artifact. Brain scans have good image quality , but other All scans have ring line artifacts . I checked xray beam and inside of the collimator but everything looks normal. what problem is this?
Has anyone experience with this kind of problem. Thanks.

Mark Adler

Log inor Register

to rate and post a comment

(268)
(20)
(167)

Marshall Shannon

Re: Picker PQ200 CT system problem

April 26, 2010 11:58

There is a PLASTIC COMPENSATOR inside the COLLIMATOR. It is broke. They just have a lifespan and yours is over. We have a few or you can find another source. To Verify remove the cover from the front of the collimator, be careful as the cover has lead on it and is a bit heavy so dont let it drop, then use a flashlight and verify a crack in the plastic compensator, then order a new one.

Good luck

Log inor Register

to rate and post a comment

(374)
(25)
(6)

Jose Morillo

Re: Picker PQ200 CT system problem

April 27, 2010 06:17

Hi Mark:
The problem are one or more bad detectors. If I understood your message, you have vertical line in all pilots exept brain pilots (Brain pilots and scans are ok), 2.- The artifact ring is in in the outside of the image (if the artifact is in a small ring in the center, then Marshall is right). The difference is due brain scan use half field and abdominal use full field image. use _dita program to find detectors. Feel free to contact me by email if you need further assistance
Jose

Log inor Register

to rate and post a comment

(10)

(1)

Sanchez Navarro

Re: Picker PQ200 CT system problem

April 29, 2010 09:51

I think you should check beam limiter selenoid,if this selenoid doesn't work correctly,you can see again ring artifacts or you should check collimator Relay board

Log inor Register

to rate and post a comment

stats
DOTmed text ad

MRI & CT Support Worldwide. Buy Parts Fast with LASO.

Comprehensive medical imaging support. We supply essential parts & maintenance for most OEMs, buy/sell quality used equipment, and offer mobile rentals globally. Contact us: info@lasoimaging.com

stats

Upcoming Auction Events

Cbrx2sales
TIMED Ending Soon
Jun 05, 2026 (157 lots)