PICKER PQ2000 ERROR 2012 PROBLEM
June 28, 2009 08:04
Hi,
We have Picker PQ 2000 Ct. System does not make rotation. We replaced Servo Module, GPU Board and PMDC board. But still no rotation. Even half turn our system gantry does not make any rotation.
When we try to run the unit we received ERROR 2012
ERP Desc. : POS MEAS FAULT
Any Commend ?
Regards,
Ercan
to rate and post a comment
Dave Adams
Re: PICKER PQ2000 ERROR 2012 PROBLEM
June 28, 2009 10:25
It seems you might be replacing a lot of boards before making the proper checks. You maybe now have more problems because you didn't make these basic checks..... It sounds to me you have an open resolver circuit to start, and who knows what now? Because of all the board swapping......
- Dave
to rate and post a comment
ORLANDO CRESPO
Re: PICKER PQ2000 ERROR 2012 PROBLEM
July 23, 2009 01:29
Your problem is the patient couch!!! Brake interlock.
to rate and post a comment
Mazher saeed
re: Re: PICKER PQ2000 ERROR 2012 PROBLEM
May 30, 2013 04:39
where is the resolver circuit?
to rate and post a comment
Doscar Ernesto Perez Pereyra
re: re: Re: PICKER PQ2000 ERROR 2012 PROBLEM
June 01, 2013 08:22
Check Brushes [ This de PCB That is turning with the rings. ] wash it.
to rate and post a comment
Mohamad Ismail
Re: PICKER PQ2000 ERROR 2012 PROBLEM
June 29, 2009 09:27
Dear Sir ,check the data brush or the microfuses located in the gantry cardfile. Your problem is a bad resolver connection.