Check the system control board leds.
You are likely to have an inverter problem, and it will be shown there.
Even if you have disabled each one of the 4 inverters at a time, you have to know what tube current can you use on your test. It will depend on which inverter is disabled.
Abdula, disconect all xsc´s cables (except power supply) and verify all test points on it
If the xsc+ it´s ok you are going measure only 1vdc at tp55, if any other tp has more than 0v you have problem with your xsc+
Specializing in GE, we keep your ICU monitors calibrated and patient-ready—fast, reliable service from expert biomed techs. Southeastern Biomedical: 888-310-7322.
Abdula Saeed El Shafeey
Picker PQ5000
November 20, 2007 03:07
Dear friends ,
For a Picker PQ5000 , I have the following fault :
Error code : 1002
XSC+ status : XSC+ interface s/w , no ALE signal . ( KV
to rate and post a comment
Anthony Guzenski
Re: Picker PQ5000
November 20, 2007 03:40
What type of generator do you have?
to rate and post a comment
Abdula Saeed El Shafeey
Re: Picker PQ5000
November 21, 2007 12:42
CRX400
to rate and post a comment
Mohamad Ismail
Re: Picker PQ5000
November 21, 2007 02:23
Dear Abdula
Check the digital control board, i faced this problem.
Regards.
to rate and post a comment
Alvaro Moreira
Re: Picker PQ5000
November 21, 2007 12:01
Check the system control board leds.
You are likely to have an inverter problem, and it will be shown there.
Even if you have disabled each one of the 4 inverters at a time, you have to know what tube current can you use on your test. It will depend on which inverter is disabled.
to rate and post a comment
Abdula Saeed El Shafeey
Re: Picker PQ5000
December 03, 2007 01:52
All warnning yellow leds on the system control board are off .
to rate and post a comment
Denis Alves
Re: Picker PQ5000
December 17, 2007 11:17
Abdula, disconect all xsc´s cables (except power supply) and verify all test points on it
If the xsc+ it´s ok you are going measure only 1vdc at tp55, if any other tp has more than 0v you have problem with your xsc+
to rate and post a comment
Abdula Saeed El Shafeey
Re: Picker PQ5000
November 23, 2007 08:49
I already tested it in another system . The digital control board is OK .
to rate and post a comment