Over 500 Total Lots Up For Auction at One Location - NJ 06/15

picking a Generator

Joseph Ladika

picking a Generator

March 24, 2004 11:19

hi
I am trying to get set up in case i have a power failure. I need to find a port generator tha will run a AC unit and a Chiller for my MRI PHILIPS T1 Magnet
Chiller 22amps 3phase 208/230
1-- compress pump
AC unit 95 amps 3phase 208/230
2-- compress pumps and a reheat unit 20kw

what size generator do i need.
thanks joe

Log inor Register

to rate and post a comment

(100)
(2)
(5)

Dave Adams

Re: picking a Generator

March 24, 2004 04:50

http://www.macgen.com/calcul.html
that site calculates power for you, 50kva Is cutting it close a 75 kva should do just fine.

Log inor Register

to rate and post a comment

Joseph Ladika

Re: picking a Generator

March 25, 2004 01:50

thanks for your help

can you tell me if the generator is a 75kva
what is kva?
is it the same a watts or Killwatts
sorry to be so dumb
can you tell me what 75kva would be in watts and killwatts
thanks again joe

Log inor Register

to rate and post a comment

(1)

Stu Dailey

Re: picking a Generator

December 15, 2004 10:39

kva is simply 65k or thousand. 65,000watts

to find what amperage you need simply apply watts law. watts divided by power source.

example. the typical MRI trailer is around 100amps @480 3phase. lets say we use a 75kva gen.

then its 75,000 divided by 480 i believe its around 156.25 amps of power.

good luck

Irei

Log inor Register

to rate and post a comment