Anwar Ahmed

Pihlips BV25 error 3 ..

January 20, 2010 09:43

this bv 25 switches ON normally but when I push the fluro or exposure button , error code 3 start appear in right lcd screen on cosole
(it is a touch console bv 25 model). and one more thing when i remove the xray tube wire (blue & brown thick wire above SE 39 card
)from its connector no error code is shown ,it seems machine work normal but when i connect those wire again problem persist
.please help....

Log inor Register

to rate and post a comment

(59)
(9)
(1)

Jorge Cid Alvarez

Re: Pihlips BV25 error 3 ..

March 01, 2010 09:19

Phlips, BV25,
Fluoroscopy check
ERROR 3
Overcurrent situation HV convertor/practix-C
1. If lamp test Ok
-Replace SE33
-Check Pactix-tank shorted
-Replace Practix-C
2.If lamp test not ok
-Check SE33 HVTD. SE35 PWRSM
-Replace SE33, 37,39
3. Check SE13 INVAL
4.Check SE13,17 CMXRSC

Log inor Register

to rate and post a comment

(9)

francisco douglas

re: Re: Pihlips BV25 error 3 ..

July 01, 2012 09:02

hi i have two uninades poblema with error 3
how to test the lampanda
  C-arm 25 bv

Log inor Register

to rate and post a comment

(177)
(20)
(3)

Chris Baklik

re: Pihlips BV25 error 3 ..

July 03, 2012 09:37

Anwar

I have the service manual for the BV25 e-mail me crisb1955@aol.com