this bv 25 switches ON normally but when I push the fluro or exposure button , error code 3 start appear in right lcd screen on cosole
(it is a touch console bv 25 model). and one more thing when i remove the xray tube wire (blue & brown thick wire above SE 39 card
)from its connector no error code is shown ,it seems machine work normal but when i connect those wire again problem persist
.please help....
Specializing in GE, we keep your ICU monitors calibrated and patient-ready—fast, reliable service from expert biomed techs. Southeastern Biomedical: 888-310-7322.
Anwar Ahmed
Pihlips BV25 error 3 ..
January 20, 2010 09:43
this bv 25 switches ON normally but when I push the fluro or exposure button , error code 3 start appear in right lcd screen on cosole
(it is a touch console bv 25 model). and one more thing when i remove the xray tube wire (blue & brown thick wire above SE 39 card
)from its connector no error code is shown ,it seems machine work normal but when i connect those wire again problem persist
.please help....
to rate and post a comment
Jorge Cid Alvarez
Re: Pihlips BV25 error 3 ..
March 01, 2010 09:19
Phlips, BV25,
Fluoroscopy check
ERROR 3
Overcurrent situation HV convertor/practix-C
1. If lamp test Ok
-Replace SE33
-Check Pactix-tank shorted
-Replace Practix-C
2.If lamp test not ok
-Check SE33 HVTD. SE35 PWRSM
-Replace SE33, 37,39
3. Check SE13 INVAL
4.Check SE13,17 CMXRSC
to rate and post a comment
francisco douglas
re: Re: Pihlips BV25 error 3 ..
July 01, 2012 09:02
hi i have two uninades poblema with error 3
how to test the lampanda
C-arm 25 bv
to rate and post a comment
Chris Baklik
re: Pihlips BV25 error 3 ..
July 03, 2012 09:37
Anwar
I have the service manual for the BV25 e-mail me crisb1955@aol.com
to rate and post a comment