Hello, I've the Planmeca Proline 2002 cc Ortho and the machine reports me the Error 57 in the display. I've turn off and then turn on the Ortho, but the error continues.
Can anyone help me?
Is the hand switch the button you press to take a radiography? I don't have this piece instaled. I only have the autoprint machine instaled. Sometimes after the error 57 appears HE 2 it seems like the
Keyboard doesn't respond to the key press. Should I connect the handswitch? How can i install it?
Thank you!!
Hi, you have problem in exposure button, praticaly is switched on. Open the exposure botton and also you can have problem in the extention cable. Put out the cable of extention exposure button and the error must deasapeare.
Send me a mail, and i send you technical manual.
Boian
Is the hand switch the button you press to take a radiography? I don't have this piece instaled. I only have the autoprint machine instaled. Sometimes after the error 57 appears HE 2 it seems like the
Keyboard doesn't respond to the key press. Should I connect the handswitch? How can i install it?
Thank you!!
HE 2 Amalgam container is 100% full (only Metasys
or Dürr).
Replace or empty the container immediately.
The suction system cannot be used before the
container is emptied or replaced. See the
User’s manual for details.
send me an email address I will send you the service manual, if your keyboard does not respond and is used to initate the exposure it probably is bad
Comprehensive medical imaging support. We supply essential parts & maintenance for most OEMs, buy/sell quality used equipment, and offer mobile rentals globally. Contact us: info@lasoimaging.com
julogo
Planmeca 2002 cc proline error
September 02, 2014 10:10
Hello, I've the Planmeca Proline 2002 cc Ortho and the machine reports me the Error 57 in the display. I've turn off and then turn on the Ortho, but the error continues.
Can anyone help me?
Thanks!
to rate and post a comment
Lenny Place
re: Planmeca 2002 cc proline error
September 03, 2014 09:01
remove the handswitch and turn unit off and on and see if it clears? could be a bad switch closed all the time
to rate and post a comment
Peter Russell
re: Planmeca 2002 cc proline error
September 03, 2014 09:01
Error 57 hand switch activated during switch on check this first
to rate and post a comment
julogo
re: re: Planmeca 2002 cc proline error
September 03, 2014 12:54
Is the hand switch the button you press to take a radiography? I don't have this piece instaled. I only have the autoprint machine instaled. Sometimes after the error 57 appears HE 2 it seems like the
Keyboard doesn't respond to the key press. Should I connect the handswitch? How can i install it?
Thank you!!
to rate and post a comment
bojan berlot
re: Planmeca 2002 cc proline error
September 03, 2014 09:02
Hi, you have problem in exposure button, praticaly is switched on. Open the exposure botton and also you can have problem in the extention cable. Put out the cable of extention exposure button and the error must deasapeare.
Send me a mail, and i send you technical manual.
Boian
to rate and post a comment
julogo
re: re: Planmeca 2002 cc proline error
September 03, 2014 12:54
Is the hand switch the button you press to take a radiography? I don't have this piece instaled. I only have the autoprint machine instaled. Sometimes after the error 57 appears HE 2 it seems like the
Keyboard doesn't respond to the key press. Should I connect the handswitch? How can i install it?
Thank you!!
to rate and post a comment
Lenny Place
re: re: re: Planmeca 2002 cc proline error
September 03, 2014 05:34
HE 2 Amalgam container is 100% full (only Metasys
or Dürr).
Replace or empty the container immediately.
The suction system cannot be used before the
container is emptied or replaced. See the
User’s manual for details.
send me an email address I will send you the service manual, if your keyboard does not respond and is used to initate the exposure it probably is bad
to rate and post a comment
bojan berlot
re: re: re: re: Planmeca 2002 cc proline error
September 04, 2014 10:19
Amalgam in x-ray?!?
to rate and post a comment
Lenny Place
re: re: re: re: re: Planmeca 2002 cc proline error
September 05, 2014 09:38
that came from the service manual
to rate and post a comment
Jose Luis Ortega
re: re: re: re: Planmeca 2002 cc proline error
March 10, 2017 10:08
Hello Lenny, i need this service manual, can you send it also to me please medicapoli@gmail.com
Thanks so much
to rate and post a comment
Lenny Place
re: re: re: re: re: Planmeca 2002 cc proline error
March 10, 2017 10:33
OK I will send
to rate and post a comment
Jose Luis Ortega
re: re: re: re: re: re: Planmeca 2002 cc proline error
March 10, 2017 04:46
Thanks a lot
to rate and post a comment