planmeca 2002

Tags: Dental, PLANMECA

issam radjah

planmeca 2002

August 30, 2015 01:24

when i try to make exposure in the begining from the expousure i get Err33, can you help me also with this?? thank you

1 person liked this
Log inor Register

to rate and post a comment

(22)
(31)

bojan berlot

re: planmeca 2002

August 31, 2015 08:50

First control the rotating part cable harnes (part 00050860), then control and clean the contacts on fuses in the tube generator bord.

1 person liked this
Log inor Register

to rate and post a comment

issam radjah

re: re: planmeca 2002

September 01, 2015 07:48

the cable and the contact on fuses in the tube generator bord work normaly
what can I do

1 person liked this
Log inor Register

to rate and post a comment

(7)
(1)

Heber Gutierrez

re: re: planmeca 2002

September 01, 2015 07:48

Hi there Bojan
I have a problem with the same equipment. This consist on the exposure is low in the panoramic even applying the technic more high and the result is the image doesn't see. But in the cephalometric the image is good. you have some idea what to happening

1 person liked this
Log inor Register

to rate and post a comment

(22)
(31)

bojan berlot

re: re: re: planmeca 2002

September 01, 2015 04:40

Replace the cassete. In the cassete is phosphorus, it loose the power.

1 person liked this
Log inor Register

to rate and post a comment