Hello Forum colleages,
When exposure preparation is initiated, there is no x-ray tube anode rotation.In the generator cabinet board d63 display 51code. also on the console desk is displayed err 119. what could be the likely faulty part in the starter unit, N81? Thanks.
check the rotor(stator) cable ohms measurement end to end. Had open wires at the tube end where it can fray from angling between table and chest positions.if you do find resistant issues then the likely cause is the cable needs to be repaired or if you have extra available in the wall throughs pull some through and cut out the bad section.
DirectMed Imaging specializes in Equipment Sales, Parts, and Support. Parts are in stock and available for immediate delivery. In-room MRI Upgrades are available now. (855) 463-3727 or sales@directmedimaging.com
Ogali Daniel
Polydoros 100 error 119
June 11, 2013 07:45
Hello Forum colleages,
When exposure preparation is initiated, there is no x-ray tube anode rotation.In the generator cabinet board d63 display 51code. also on the console desk is displayed err 119. what could be the likely faulty part in the starter unit, N81? Thanks.
to rate and post a comment
Lenny Place
re: Polydoros 100 error 119
June 11, 2013 06:16
check your power supplies on the d61 board
to rate and post a comment
R GOPAL
re: Polydoros 100 error 119
June 12, 2013 07:50
Check power supplies on N81.
to rate and post a comment
ErnestoQ
re: Polydoros 100 error 119
June 12, 2013 11:11
check the rotor(stator) cable ohms measurement end to end. Had open wires at the tube end where it can fray from angling between table and chest positions.if you do find resistant issues then the likely cause is the cable needs to be repaired or if you have extra available in the wall throughs pull some through and cut out the bad section.
to rate and post a comment
ErnestoQ
re: Polydoros 100 error 119
June 12, 2013 11:11
do an ohm check of the rotor(stator) cable end to end.
to rate and post a comment
Pandoros
re: Polydoros 100 error 119
June 13, 2013 12:15
The usual suspects include stator cable, x-ray tube, inverter
to rate and post a comment