Polydoros 100 error 119

(20)
(2)

Ogali Daniel

Polydoros 100 error 119

June 11, 2013 07:45

Hello Forum colleages,
When exposure preparation is initiated, there is no x-ray tube anode rotation.In the generator cabinet board d63 display 51code. also on the console desk is displayed err 119. what could be the likely faulty part in the starter unit, N81? Thanks.

1 person liked this
Log inor Register

to rate and post a comment

(2681)
(673)
(21)

Lenny Place

re: Polydoros 100 error 119

June 11, 2013 06:16

check your power supplies on the d61 board

Log inor Register

to rate and post a comment

(78)
(7)

R GOPAL

re: Polydoros 100 error 119

June 12, 2013 07:50

Check power supplies on N81.

Log inor Register

to rate and post a comment

ErnestoQ

re: Polydoros 100 error 119

June 12, 2013 11:11

check the rotor(stator) cable ohms measurement end to end. Had open wires at the tube end where it can fray from angling between table and chest positions.if you do find resistant issues then the likely cause is the cable needs to be repaired or if you have extra available in the wall throughs pull some through and cut out the bad section.

1 person liked this
Log inor Register

to rate and post a comment

ErnestoQ

re: Polydoros 100 error 119

June 12, 2013 11:11

do an ohm check of the rotor(stator) cable end to end.

Log inor Register

to rate and post a comment

Pandoros

re: Polydoros 100 error 119

June 13, 2013 12:15

The usual suspects include stator cable, x-ray tube, inverter

Log inor Register