Some body,
about Error 037 polydoros IT,
some body solve this trouble?
I have a this error, I changed tube RX with a similar model and then I test HV transformer clean the contact cable and fill oil inside.
i try to warn up tube on servide mode and start this the system not allowed and box dialogue appear on screan "COMUNICATION ERROR - PLASE OPEN DIALOG AGAIN" MYPOLYD
this box appear ever time to try warn up o learn filament.
i want to change D500 because not find other error on system
I hope yours kindly reply
thanks
Err037: Delta_kV 0
Description:
• Difference in value between positive and negative high voltage side is too great
Possible Causes of Error:
• Short circuit on one side in the high voltage circuit
• Strong non-symmetry in the high voltage transformer
• Defective measurement sensing on one side
Corrective Measures:
• Check whether the difference in high voltage is less than 20%; measure between kV Plus and kV Minus on D500_X186 at Pin P_UT and Pin N_UT:
- if the difference is greater than 20%
-> replace tube unit or high voltage transformer, but first read out/note the tube data
- If there is no difference -> replace D500
BMI-PMI services nearly every make, model, and modality through our nationwide team of nearly 200 local service engineers. Flexible contracts, 24/7 remote monitoring, and ISO 13485 certification. www.brownsmedicalimaging.com / 800-701-9729
Germni
Polydoros IT Error 037
September 04, 2019 09:33
Some body,
about Error 037 polydoros IT,
some body solve this trouble?
I have a this error, I changed tube RX with a similar model and then I test HV transformer clean the contact cable and fill oil inside.
i try to warn up tube on servide mode and start this the system not allowed and box dialogue appear on screan "COMUNICATION ERROR - PLASE OPEN DIALOG AGAIN" MYPOLYD
this box appear ever time to try warn up o learn filament.
i want to change D500 because not find other error on system
I hope yours kindly reply
thanks
to rate and post a comment
Konstantin Kuznetsov
re: Polydoros IT Error 037
September 06, 2019 09:12
Err037: Delta_kV 0
Description:
• Difference in value between positive and negative high voltage side is too great
Possible Causes of Error:
• Short circuit on one side in the high voltage circuit
• Strong non-symmetry in the high voltage transformer
• Defective measurement sensing on one side
Corrective Measures:
• Check whether the difference in high voltage is less than 20%; measure between kV Plus and kV Minus on D500_X186 at Pin P_UT and Pin N_UT:
- if the difference is greater than 20%
-> replace tube unit or high voltage transformer, but first read out/note the tube data
- If there is no difference -> replace D500
to rate and post a comment